Hi Bryan
if you have the network analyser reference manuals at hand you should be able to discover
what the actualvalues are.
YOU SHOULD NOT GENERATE YOU OWN USER VALUES .
The instrument is calibrated to do the operation for you.I suggest you read further regarding calibration of
network analysers. They are dedicated instruments to provide Readings such as S parameters,insertion loss,
gain, isolation.
What you can do is generate calibration files for different setups using the same HF cables connected to different
DUTs.
For example A full 2 port calibrate entails you using a 50 ohm load, open cicuit and short circuit reference on Port1
output( point were the DUT is connected) this is then repeated for Port
2 ( connection were DUt output connects to return cable. If I remember
correctly you actually compensate for 12 error algorithms!!
Remember to allow the instrument to stablise before doing thia calibration switch on for at least 30 minutes before hand.
Also IF YOU CHANGE THE connection cables between the DUT and network analyser then recalibrate.DO NOT USE OLD CALIBRATION FILE
Changing the cables effects the reflection coefficients making any reading uselessif you havn't recalibrated.
chow
xseadog