Communication stucks with Ketihley 2700 data logger
I've been facing a technical issue with Keithley 2700. Communication lost occurs after hours of data acquisition:
Even if i stop the .vi and restert it I cannot have communication with Keithley 2700. I must restart my PC all the time in order to the communication be fine.
I use 1m long RS-232 cable.
Have you got any ideas please?
Kind regards,
Balázs
Message 1 of 8
(4,419 Views)
Hi Balazs,
do you get an error after more than 9½ hours?
Message 2 of 8
(4,399 Views)
Showing some code *might* help.
Are you using a USB-RS232 converter or a straight RS232 cable with a dedicated serial port on a PC?  If USB is involved look at your PC power settings and disable the USB selective suspend setting - https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P7POSA0&l=en-CA
Craig
Message 3 of 8
(4,360 Views)
I apologize for the late reply.
Let me answer your questions:
- Error code 85 can occur any time and pops up (refers to) the following VI:
Error 85Error 85
When i stop the data acquisition after error 85, i get another error code as :-1073807339
- I am using straight RS232 cable with the lenght of  about 1m. Note: Error 85 occured also when i was using USB to RS232 adapter.
I attached the simplified VI that i use in my project.
Can you find some bug to be fixed in my code?
Kind regards,
Balázs
Message 4 of 8
(4,303 Views)
Hi LTBalazs,
error 85 is from ScanFromString: the received string doesn't fit to the scan format string…
All you need to do is to handle the error: right now there is no error handling at all!
This is a TimeOut error, probably due to not handling the serial port correctly…
The loop of your VI can be simplified like this:
What's the point of calling the Close (and Error handler) function after the loop when the loop never will stop?
Use a boolean button instead of this FALSE constant!
Message 5 of 8
(4,284 Views)
Hi GerdW,
After the failed communication with Keithley 2700, I stopped the vi and started it again. I then got the following error:
Error code: -1073807339Error code: -1073807339
This is something that cannot be solved by restarting the vi or quitting labview and powering off/on Keithley.
I have to restart both: the PC and Keithley to start data acquisition with keithley again.
Do you think i can manage this phenomenon in labview in run time? How?
Kind regards,
Balázs
Message 6 of 8
(4,253 Views)
Hi Balázs,
which kind of RS232 port do you use?
Directly on PC motherboard, PCI(e) expansion card, USB-to-RS232 converter?
Message 7 of 8
(4,251 Views)
Hi GerdW,
It is RS232 connection on PC motherboard.
I tried to plug the rs232 cable to another RS232 socket, but error still occured:
'Error -1073807339 occurred at VISA Read in Keithley 27XX.lvlib:Error Query.vi->Keithley 27XX.lvlib:Error Query (Multiple).vi->Keithley 27XX.lvlib:Configure Scan.vi'
Then plugged the rs232 connector back to its original socket. Result: Connection is OK.
It seems to me that the connection 'hangs' after some time.
Kind regards,
Balázs
Message 8 of 8
(4,206 Views)
