DAQ assistant - error when applying a calibration

Hello,

My situation is simple. I get the pressure from a pressure sensor.

It returns 0...10 V for a -1...6 bar pressure

 

I just open the DAQ assistant (input => voltage) and returns the value. I set Continuous Values because I want to have the continous values. I set 100 Hz because I don't need so much values every second (and also 1000 Hz bug).

 

It works 100% fine.

 

Problem is : it's not calibrated perfectly. So I use the calibration. I have small differences. I attached the picture. But when I use the calibration I have the error (image attached). If I unclick "enable calibration" (meaning desactive it), it works.

 

I set from 0 to 10 V, what my sensor delivers.

 

Any idea what this is ? "specified value is smaller than the minimum value" "corresponding value : 10"

doesn't make any sense to me.... thanks so much if someone can help !!

 

faaa.jpeg

error.jpeg

DAQ.jpeg

  

0 Kudos
Message 1 of 2
(1,274 Views)

Your Uncalibrated value for the Reference of 8.8 is 10.126.  That is higher than 10.