DAQmx write Analog DBL 1Chan 1Samp - hi/low output NI USB-6009 - also error code -200077

Solved!
Go to solution

I'm attempting to get 5V high 0V low ouput from ao0 of a NI USB-6009 based on a switch in LabVIEW using DAQmx write Analog DBL 1Chan 1Samp.

 

When using a voltmeter connected to ao0 and analog-ground on NI USB-6009; Im not getting anything.

 

I'm also getting error code -200077 for some reason:

DAQmx Start Task.vi:4<append>
<B>Property: </B>AO.Max
<B>Requested Value: </B> 10,0
<B>Valid Values Begin with: </B> 0,0
<B>Valid Values End with: </B> 5,0

<B>Channel Name: </B>Dev1/ao0

<B>Task Name: </B>_unnamedTask<28425>

Download All
0 Kudos
Message 1 of 3
(3,980 Views)

Found this thread with the same problem:

http://forums.ni.com/t5/LabVIEW/Generatiing-Analog-output-Voltage-for-a-Servo-Fan/td-p/989754

 

Tried to do the same, but it's still not working.

 

 

 

0 Kudos
Message 2 of 3
(3,955 Views)
Solution
Accepted by topic author Thorstif

Set the output terminal configuration to RSE instead of default.

 

Set the maximum value to five and the minimum value to zero.

 

See attachment.

0 Kudos
Message 3 of 3