DC Fail while setting the voltage for SM15K delta elektronik power supply

Solved!
Go to solution

I am facing a DC fail(blinking) error in the hardware which blinks while setting the voltage. i tried to ramp up the voltage but it doesn't solved the problem. How to solve this error?

RepMat_0-1762859429028.png

RepMat_1-1762859456899.png

unforutnately i cannot add the full vi but i attached a screenshot of it. 

0 Kudos
Message 1 of 5
(116 Views)

Were you able to find from the manual what that DC fail symbol means and what causes it?

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 2 of 5
(92 Views)

DC FAIL - This indicator is active if the DC power terminals are 5% below or above the set value for the voltage.

 

I tried to set the voltage through front panel at different speeds and is not able to reproduce it. however through the LabVIEW example i could reproduce and not able to set the voltage due to DC fail error

 

i used different ramp settings to eliminate this 5% requirement but still facing the error.

Download All
0 Kudos
Message 3 of 5
(69 Views)

@Johnzahey wrote:

DC FAIL - This indicator is active if the DC power terminals are 5% below or above the set value for the voltage.

 

I tried to set the voltage through front panel at different speeds and is not able to reproduce it. however through the LabVIEW example i could reproduce and not able to set the voltage due to DC fail error

 

i used different ramp settings to eliminate this 5% requirement but still facing the error.


I see it as; the instrument does something automatically which you're not programming it from LabVIEW, most common scenario is that the instrument may be continuously adjusting the ranges or limits but you may be operating in the same limit or not changing the limit in the correct order before sourcing.

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 4 of 5
(56 Views)
Solution
Accepted by topic author Johnzahey

i changed the program source of power to front panel and is working now

0 Kudos
Message 5 of 5