Thank you for Your answer!
I use a nullmodem cable, and I get response on the *IDN? query.
I use the HP34970A sample read thermocouple temp.vi.
First i go the errormessage 205 on the Agilent instrument.
225: Settings conflict: DMM disabled or missing
This command is valid only when the internal DMM is installed and
enabled. Use the INSTrument:DMM? command to determine the state of
the internal DMM.
Should the internal DMM of the instrument be turned on?
I tried to turn it on and i got error 305: Not able to perform requested operation
The requested operation is not valid for the specified channel. You may
have tried to a configure a channel for current measurements (valid
only on channels 21 and 22 on the 34901A module). Or you may have
tried to configure scaling on a module that does not connect to the
internal DMM.
I am trying (at first) to read temperature throug a 4W RTD, and I configure channel
101 (who correctly pairs with n+10=111).
I use 19200 baud, 8-bit, no parity for the COM settings.
-A