03-11-2008 12:24 PM
03-11-2008
01:04 PM
- last edited on
05-27-2026
07:57 AM
by
Content Cleaner
You may want to start here.
There are also lots of examples that ship with LabVIEW, and it also has a DAQ Assistant that can create code for you.
03-12-2008 11:52 AM
03-12-2008
12:56 PM
- last edited on
05-27-2026
07:58 AM
by
Content Cleaner
Without knowing more about the way you've set it up, and exactly what "weird results" means, I can only speculate. You may want to look over these documents in terms of the excitation voltage: [links below removed; articles no longer available]
The Excitation Voltage from My SCXI-1520 Does Not Match Its Configured Value
How Can I Test the Voltage Excitation Level of a SCXI-1520?
Reading a Voltage Signal With the SCXI-1520
03-13-2008 07:04 AM
03-13-2008 09:25 AM
03-13-2008 09:57 AM
03-13-2008 12:12 PM
Hi Akash Kumar,
Fisrtly, I would like you to verify the way you wired your sensors. Are you really in differential mode?
Secondly, could you explain precisely what you are using as sensors and what kind of measure you want to do.
I have changed your VI even if it has the same effect. You don't need to put a read task function after a DAQ assistant which already do it.
What is the purpose of the subVI?
Regards,
03-13-2008 12:39 PM
03-17-2008 05:26 AM
Hi Akash Kumar,
Thank you for these informations. If you are sure that your are in differential mode, the other channels have no effect. I advise you to make the calibration of your SCXI 1520 module. Secondly, may you verifiy in your task configuration the range of input you set? Decreasing this setting will increase your amplification. And you can change the cut-off frequency of the low-pass filter to eliminate noise.
May you try to verify those parameters and make the calibration of you module?
Regards,