Doubt about the sampling frequency when generating and acquiring
when generating and acquiring samples, the maximum sample rate is the maximum board sample rate divided between the number of generating and acquiring task?
thanks
(2,859 Views)
Hi hcobas,
- usually AI and AO tasks are separated, not interfering each other.
- max sample rate is determined by your hardware. Is it using a MUX? Does it support hardware timing? Hardware timing for all I/O channels?
(2,854 Views)
Hello GerdW, thank you for your reply.
I am using the DAQPad6016 acquisition board which have a maximum sampling frequency of 200 kS/s. Initially i want to generate a excitation signal and acquire the response signal. So i am using an analog output for the waveform generation and analog input for the waveform aquisition. So in these terms i am not using the mux because as you said they are separated task.
so these means that i can generate and acquire to the maximum sampling frequency (200 kS/s) for each one of them?
thank once more time.
Houari
(2,846 Views)
Accepted by topic author hcobas
Hi Houari,
you should read the specs of your DAQpad more carefully!
It clearly says: Analog inputs = 200kS/s samplerate, but analog outputs = 300 S/s sample rate or just even 50S/s for hardware timing!
(2,837 Views)
Thank you very much. I really did not perceive it. I see the hardware wil not work well enough.
Thank you very much, really
Houari
(2,834 Views)
