Don't know what a CRO is but I will try my best to answer your question. Two things to consider when acquireing an ECG signal: 1. Resolution and 2. Sample Rate.
1. The output from standard ECG leads is in the milliVolt range (~1 to 3 if i remember correctly). The smallest voltage change the average NI data acquisition board will be able to detect is ~1 milliVolt. So, you will have to amplify and filter the signal from the leads to have a good enough signal to visualize in your VI.
2. The peak Hz in a ECG signal, the QRS complex, is around 17 Hz (again, my memory is a little fuzzy here) so, theoretically, you can sample data around ~34 samples/sec and get away with it. However, I would recommend a x10 sampling rate (~170 Hz). 300 sample
s/sec should be fine but don't oversample or you will introduce unwanted noise in your signal.
So, I think the VI you are using to display the ECQ signal should be fine. But, I think an emphasis should be placed on how you acquire the signal.
Hope this helps.