Hi Solarman,
Thank you for posting on the NI Forums.
I have taken a look into your issue and it would seem that several customer's have seen this issue before, but the issue often seems not to be reproduceable.
In all cases I could find where this issue has occured under LV 7.x (as opposed to 6.x) the issue has been resolved by an uninstall of NI software, then a registry clean and then a reinstall. Please see the attached .doc for methodology of this.
If you could post your scf file to this forum then I can verify if your scf is indeed corrupt. Many thanks,
Rob
Message Edited by RobS (AE) on 05-22-2006 11:15 AM
05-23-2006 05:48 AM
05-23-2006 05:51 AM
Morning Paul,
I'll take a look at your .scf file for you today and let you know what I find.
Thanks,
Rob
05-25-2006 03:55 AM
Rob,
On reverting to V7.0 I found something which may or may not have something to do with the error but I thought I should share it with you. The reason for the edits to the .scf file was to add tags for an FP-DI-330 module which I recently installed (labelled OTR1.....SMT 111.... to 114....). In V7.1 I had been able to use these tags but when I reverted to v7.0 I found they didnt work and when I looked at them using the tag monitor they showed up as "access denied" under status. When I had a closer look at the configuration they had been set as Input/Output instead of Input only. Setting them to input only allowed access to be restored and the application now runs fine.
The fact that I could access the tags in v7.1 when they were not correctly configured suggests that something changed between versions.
Regards
Paul