Background description of system:
The problem:
A few days ago, I could run the program shown in the JPG perfectly fine. Now, I am always obtaining these generic file i/o errors. I used to be able to output files without a hitch, but now the program is always stuck in the loop attempting to get a measurement. I have already looked at potential timing issues, hard drive space issues, and path name length issues, but none of those seem to be the problem. Between now and the time when the program was working perfectly fine, no changes were made to the computer besides creating more .txt files for data.
I'm at a loss as to what to do right now. If there is anything else I could provide that you think would help, let me know and I will do that as soon as possible.
Thanks for reading!
Solved! Go to Solution.
08-22-2013 01:08 PM
Look at that error window a little closer. You will see that the GPIB I/O was aborted. This means that the communication with the instrument is having issues. Common problems here include 1) using the wrong address, 2) the instrument not configured to talk over GPIB, or 3) the instrument is not hooked up the the GPIB bus. First make sure you can communicate with the instrument through MAX and troubleshoot from there.
08-22-2013 02:08 PM - edited 08-22-2013 02:15 PM
I just went in and made sure that the 2182A had the correct address and allowed GPIB control (both of which were yes). I also made sure the the GPIB cables were in there nice and tight. Everything seems to be working well now, so I think it was the cables themselves. I think this issue has been solved for now. If it happens again, then I will return so that the problem could be further diagnosed.
Thank you! And kudos to you...