LabVIEW

cancel
Showing results for 
Search instead for 
Did you mean: 

How Do I Create a Single Delayed Output Pulse (Analog Voltage) From An Input Reference Trigger (Analog Voltage)?

Solved!
Go to solution

Hi I am working with a combustion chamber and using a DAQ system (with NI SCB-68 hardware) to read the pressure within the cylinder (as analog voltage). I am trying to output a delayed, 1-millisecond, pulse at 5 volts once the pressure within the cylinder has risen above 5 bar (which corresponds to an analog voltage of 0.25 V). I also want to record 30 ms of samples before the trigger and 220 ms of samples after the trigger. The following image visually shows what I'm talking about.Pulse Diagram.JPG

I have created a LabVIEW VI (which is attached) but I keep running into 2 problems:

  1. When I run it with Finite Samples after a period of time I get Error-200281which I don't entirely understand.
  2. Using the "Elapsed Time" Express VI to keep time for the pulse I cannot get a resolution of 1 millisecond, so the pulse is not generated when I set the window between 0.003 and 0.004 seconds for the pulse to be high (i.e. the resolution of "Elapsed Time" seems to be too coarse).

I am a beginner at LabVIEW so sorry if my questions are trivial or if my VI makes no sense but I have been stuck on this for more than a week. Any help would be greatly appreciated!

 

Thank you,

Morgen

0 Kudos
Message 1 of 7
(4,814 Views)

Here's an explanation of the error code -200281:

 

https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019NNASA2&l=en-US

 

As far as getting a 1ms pulse inside that While Loop, I'm betting that is takes mroe than 1ms to iterate. The Elapsed Time VI can handle a 1ms increment, but you've got a lot going on in the While Loop, the Write to Measurement File VI is opening, writing, and closing the file each iteration.

Message 2 of 7
(4,765 Views)

Thank you so much for your help! The explanation of the error code is extremely helpful and that makes sense why I am unable to get the 1ms pulse. Do you have any suggestions on how I mght be able to produce the pulse without using such a complicated While Loop? I really appreciate your help.

 

Thank you,

Morgen

0 Kudos
Message 3 of 7
(4,756 Views)

What device are you using for your AO?

Message 4 of 7
(4,741 Views)

I'm using an NI PCI-6259 DAQ card with an SCB-68 interface block. Once again, thank you so much for your help. I really appreciate your input.

 

Mrogen

0 Kudos
Message 5 of 7
(4,735 Views)
Solution
Accepted by topic author Morgen

This is not a good way to trigger a pulse.Block Diagram with Annotation.jpg
Use a DAQmx trigger to send the pulse when your acquired signal rises above 250mV like you specified.

 

See this for DAQmx Triggering:

https://www.ni.com/en/support/documentation/supplemental/06/learn-10-functions-in-ni-daqmx-and-handl...

Message 6 of 7
(4,715 Views)

Thank you so much for the link! This is extremely helpful. I knew my VI was really inefficient but I wasn't sure how to clean it up. Also you were right before about the loop taking  too long to iterate. Even after I moved the "Write to Measurement" outside the loop it was still taking more than 17 ms to iterate.

 

Thank you,

Morgen

0 Kudos
Message 7 of 7
(4,707 Views)