Well, you can't get hardware-timed DIO with the 6110, but you can probably use its counters to capture the timestamp you need.
One method would be to set up a buffered edge-counting task. The analog input's sample clock would be specified as the source signal. This will cause the counter's value to increment at the beginning of each sampling cycle, acting as a measure of time from the start of analog input.
The falling edge of the relay signal would be the counter's gating signal. For each instance that the relay disconnects, the count value ("time") at that instant will be stored in the counter's buffer for subsequent retrieval.
Note one small timing subtlety: while the very first sample clock edge puts analog data into the 0th array inde
x, it will also be incrementing the counter time to 1. This off-by-one deviation will persist.
There are other variations on this theme, depending on whether you also need to capture re-connection times, or whether there is always one and only one disconnect of interest, etc. You can likely get just the behavior you need using one or both of the on-board counters.
-Kevin P.