How many sample rate groups I can use in a 9189 chasis with 8 NI 9231
I have a 9189 cDAQ chassis and several NI 9231 cDAQ modules.
How many AI tasks with different sample rates can I have.
Thanks.
Ian
(3,275 Views)
Accepted by topic author TulaRen
3
(3,260 Views)
Thanks. That's what I remembered, but I just cannot find anything reference to this from ni.com.
Ian
(3,256 Views)
Its down deep in a cDAQ whitepaper
(3,246 Views)
Hello Ian,
In this case you can actually only run 2 hardware timed analog input tasks with different sample rates because the 9231 is a DSA module. This is because the cDAQ 9189 is a Gen II chassis and Gen II chassis support a maximum of 2 synchronization pulses. This is called out here.
Conrad
C Series IO R&D
(3,238 Views)
Thank you for the correction and link
(3,231 Views)