First you need to buy a power supply for the pressure transducer, probably
a 24 Vdc supply.  The nidaq card can only measure voltage from -10 to +10.
So you need to convert the current output on the transducer to a voltage.
Connect a resistor from one lead to another and use the nidaq card to
read the voltage across the resistor.  Use a resistor that maximizes your
resolution.  Since you have a 4-20mA transducer, a 500 ohm resistor
will read from 2-10V, and a 250 ohm resistor will read from 1-5V.
The larger the range the better.
By the way, if you haven't bought it yet, you will need a breakout
box and cable connector for the card.  I use the BNC 2090 breakout
box and two cables (183569B-01 and 182419B-01) to connect to my
PCI-MIO-16E-4 card.
Boozz wrote:
> Hi,
> I (actually my boss, but who cares) bought a NI-DAQ-card type: PCI-MIO-16E-4
> from NI. I want to use it in combination with LabView. What's the best
> connection to get some good data from this transmitter.
> By the way, I'm a mechanical Engineer, so please be carefull with me when
> you're talking electronics...
> Transmitter is:
> Endress+Hauser
> it needs a 11..30 VDC power supply,
> 2-wire type, 4..20mA output.
> Do I have to use an external power supply? How do I best connect these
> things?
> To be more precise: I'd like to connect 3 of them...
> Thanks,
> Ronald