IV measurement example for PCI-MIO-16XE-10

Is there Labview IV measurement examples for PCI-MIO-16XE-10 board ? I'm especially interested in producing the clean voltage out from the board. I'm seeing 10 meV noise in the analog voltage output.
0 Kudos
Message 1 of 3
(2,564 Views)
The shipping examples that go along with LV for analog input and output should work with your board without a problem. Furthermore, you should check the analog output channels in MAX, by using the Test Panel Resource (Analog Output) of the board and generating either a DC or Sine Wave voltage, which you can read back in the AI channels. If you still read noise out of the AO channels directly from MAX, you should contact NI tech support.

Hope this helps.

Aguila.
0 Kudos
Message 2 of 3
(2,564 Views)
Hello,

Reading the generated voltage with one of the anlaog input channels is the most recommended way to see if your board is giving an offset.

I would also recommend to run the E-Series Calibrate VI. You can find this VI in the LV Functions Palette: Data Acquisition>>Calibration and Configuration. In the following link you will find an example that shows how to do this.

Calibrate and E-Series Board using Internal Voltage Reference

Good luck with your application!
0 Kudos
Message 3 of 3