05-10-2021 12:37 PM
Good Afternoon!
I am having a bit of trouble connecting to a Signal Recovery 7280 Lock-in Amplifier from Labview. I am really not sure of what the error could be. The amplifier is viewable in NI MAX, the *IDN? command does not work, however, I can send other commands (such as VER to retrieve the version of the amplifier software) through MAX and receive a response no problem. When running the program, however, I run into a "-1223 Error."
Any help or guidance is greatly appreciatted!
05-11-2021 10:42 PM
Hi,
1. By "*IDN? command does not work" do you mean there is no respond at all after entering the command, not even an error display? . If it is that case, the reason might be that your instrument is not SCPI compliant. More detailed explanation can be found here.
2. I believe 7280 Lock-in Amplifier is a third party devie and not manufactured by NI. The error -1223 means that your driver is not intended for this instrument, you need to download a suitable driver for your instrument or develop one yourself. Please refer to the below links for more information. I hope this helps
Developing LabVIEW Plug and Play Instrument Drivers
Fiona, NI
05-12-2021 02:17 AM
Hi Yacob,
did you read page 3 of the datasheet?
Which device driver do you use? Which port of the device do you use?