12-22-2005 12:45 PM
12-22-2005 10:42 PM
Hi Marc,
Which VI are you using? Are you using any of example VI's shipped with Labview?
and what is the Hi low limit/Range/Gain settings you are using for that channel?
Regards
Dev
12-23-2005 08:03 AM
12-23-2005 09:05 AM
12-23-2005 09:10 AM
12-23-2005 07:21 PM - edited 12-23-2005 07:21 PM
Hi Marc,
Is MAX sampling at 300KHz (each channel) when you're watching the signal there?
I've observed "parasitic-capacitance" at the A/D converter which (like Lynn said) is switched between analog-inputs on E-series devices. It's easy to stumble across the demonstration: while setting-up a multi-channel continuous DAQ (say 1KHz), leaving one channel unconnnected, the value measured on the "open" channel tracks the signal on the channel sampled previously. The hard part is convincing your supervisor that this "coupling" is OK. ![]()
I'd have suggested adding an "empty-channel" tied to ground, immediately-prior to the signal being discussed, but it sounds like you wouldn't be able to sample more channels fast enough. Just copied the following out of the "E-series Help" text, in case it my be of use:
"When the multiplexer switches from channel 0 to channel 1, the input to the PGIA switches from 4 V to 1 mV. The approximately 4 V step from 4 V to 1 mV is 4,000% of the new full-scale range. For a 12-bit device to settle within 0.012% (120 ppm or 1/2 LSB) of the 100 mV full-scale range on channel 1, the input circuitry must settle to within 0.0003% (3 ppm or 1/80 LSB) of the 4 V step. Some devices can take as long as 100 µs for the circuitry to settle this much."
Cheers!
Message Edited by Dynamik on 12-23-2005 07:24 PM
Message Edited by Dynamik on 12-23-2005 07:26 PM
01-05-2006 12:56 PM
Ok, so here's my next question. We're looking at getting an S Series card for this test, so that we can have a higher sampling rate on each channel. Since the channels are independent and sample simultaneously, the capacitance on the channels shouldn't be a problem, right? I'm sure there is still some crosstalk, but I'm hoping this would solve the capacitance problem and also give us a higher sampling rate.
Thanks for the help.
01-05-2006 01:51 PM
01-06-2006
10:35 AM
- last edited on
12-02-2025
10:16 AM
by
Content Cleaner
Hello Marc A,
From what you're describing about your readings, it sounds like you are experiencing the results of what we call 'ghosting' or 'cross talk', i.e. rapidly multiplexing a wide range of analog input signals from multiple channels through a single ADC (analog to digital converter) results in the amplifier on the card not being given enough time to settle to the exact value, resulting in bad readings. The reason you are seeing different data in MAX and in LabVIEW might be based on what DAQ driver you are using in each environment (NI-DAQmx or Traditional NI-DAQ). One of the improvements built into the NI-DAQmx driver is that when the user specifies a sampling rate for multichannel analog input tasks, the driver automatically tries to space out the samples so as to give the amplifiers on the card the maximum time within that sampling rate to settle to their correct value. In the case of the Traditional NI-DAQ driver, the samples are taken from all the channels as fast as possible, resulting in less settling time on multichannel acquisition. So, the first thing I would recommend is to rewrite your LabVIEW program using the NI-DAQmx VIs in order to take advantage of this performance improvement.
As for your question about S-series cards, the simultaneous sampling functionality equates to each channel on these cards having a dedicated ADC and amplifier. Because multiplexing a range of samples through a single amplifier is not performed on these cards, you should definitely see an improvement in the amount of ghosting in your signals. Here is a link to a great Application Note that discusses the concept of amplifier settling times in more detail:
Is Your Data Inaccurate Because of Instrumentation Amplifier Settling Time?
I hope this helps!
Regards,
Travis G.
Applications Engineering
National Instruments