Making PXI-4060 and PXI-4065 hardware compatible

We have got problem when been forced to substitue the PXI-4060 DMM with the newer PXI-4065.
(4060 is going out of market)
The problem is hardware related,  (software works fine WinXP/LabView7.1)
 
Both 4060 and 4065 has 4 connectors  Hi, Lo, Sense+  and Sense -
There is no compability problem when measuring Volt or resistance,
but the problem is for current measurements.
Thats because the newer 4065, must measure current between the Lo and Sense+
connectors (for 4060 between Sense+ and Sense-)
See attached picture for better understanding the hardware problem between 4060 and 4065
 
To make the problem clearer, we uses 2 SCXI chassis with multiplexors, switches, ....
Earlier had we 2 common mux lines going to the 4060  (one to Hi and Lo, the other one to
Sense+ and Sense-).
But for the 4065 DMM, we need 3 mux lines,  (the 2 as in the 4060 case and now a mux line
to Lo and Sense-)
 
Now for the question.
We are not able to borrow mux capacity from the SCXI, so the solution is to buy a extra PXI relay board.
We want to maintain only 2 mux lines up to DMM in the system,  we need therefor to relay the Sense- connector to
Lo connector when we are going to measure current.
The question is:  Is it allowed to connect Lo connector and Sense- connector with a relay, when measuring current?
                            Or will the DMM be damaged inside, when Sense- and Lo are connected at current measuring?
 
 
 
 
 
0 Kudos
Message 1 of 2
(2,924 Views)

Hi Nilsta,

I know you had been working with Andreas on this issue and I think I can clear up your question here.  The NI-4065 has internal relays that connect/disconnect based on the type of measurement you are trying to take. So having the Sense- still connected does not cause any harm to your DMM.  The current all goes to the correct lines.  I actually tested this setup and confrimed the application first hand to confirm that measurement values would not be incorrect. Hope this clears up the question.

Knowledgebase: Do I Need to Disconnect the Voltage Source When I Make a Current Measurement on a DMM...

Alex A.
Applications Engineer
0 Kudos
Message 2 of 2
(2,871 Views)