Multiple sample clock pulses were detected within one period of the input signal error
What do you mean? I want to count the rising edges, of CH A, between one period of the index. I don't need to count the index. The encoder I am using is 4PPR, and there are 4 Ch A rising edges during a period of the index.
CH A is yellow, and the index is blue.
Message 11 of 19
(1,212 Views)
Is the blue line measured or calculated?
Tim
GHSP
Message 12 of 19
(1,208 Views)
That image was from a scope. I measured both signals, to ensure they were correct, and just screen grabbed the scope screen to post it here.
Message 13 of 19
(1,206 Views)
This one https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P8g1SAC&l=en-US says
2. Also there cannot be two sample clock pulses within one cycle of the input signal. To avoid this you can set the sample clock frequency to half of the input signal frequency.
I think it means timeout waiting for clock pulse. Try to reduce the frequency you set in Timing VI
Also check if this applies
https://forums.ni.com/t5/Counter-Timer/PFI-channel-crosstalk-Error-Code-201314/td-p/3043909
Solution there was to add a digital filter because sometimes it registered a short noise clock pulse.
Message 14 of 19
(1,205 Views)
Oops, queued this up for a reply, then came back to it a couple hours later without refreshing to see that there's been a lot of conversation in the meantime.  Oh well, here it is anyway...
The error you're getting is happening at the hardware level and isn't *caused* by your code.  My guess is that the index pulse signal isn't clean and a single clock edge manages to register as multiple edges in rapid succession.
Dunno your device, but you may be able to use DAQmx property nodes to do digital filtering on that signal.  So you may be able to *resolve* your problem using additional code.
-Kevin P
Message 15 of 19
(1,200 Views)
Tim
GHSP
Message 16 of 19
(1,195 Views)
Let me recommend some changes for the sake of diagnosing and troubleshooting this issue.
Try capturing samples for dozens of revolutions.  Find the count differences for every successive pair of samples and post the result here.  Let's start getting a better sense of how regular or irregular things look.  I see no obvious reason why you'd get count differences of 1 or 2.
-Kevin P
Message 17 of 19
(1,193 Views)
The snipped you posted appears to count the A edges precisely, but does not have a stop for the rising edges of the index.
Kevin:
I had the loop run 25 times, and captured a few runs results shown below.
Message 18 of 19
(1,179 Views)
Ok, that data is *nothing* at all like what it should be.  You don't have a small intermittent problem, you have a deeply systematic problem.
Given the scope trace and your description, you *should* see count values that increment by 4 with each sample.  Given the lack of sync between encoder and "index" pulse, I wouldn't be alarmed by an occasional increment of 3 or 5.
The runs of 0 values are almost certainly caused by getting a task error (perhaps the one you started the thread with), so those "make sense" even while helping to indicate trouble.
All the other values and increments do *not* make sense.
I kinda suspect you may not have your digital signal grounds referenced together.  Or there's some other signal incompatibility between your encoder, motor, and DAQ device.  Your USB-6341 will want single-ended digital logic.  TTL will be supported, perhaps also CMOS.  You can run into weird problems if you directly wire differential sensor outputs to a single-ended digital DAQ device.
-Kevin P
Message 19 of 19
(1,173 Views)