NI-9237 analog input channels.  Test Panel Input voltage range is -25m to +25m volts.
Can this board allow higher voltages, 5 volts?
If wrong hardware for my application, what DAQ board?
If wrong board, then a sensor in mV range for reading pressure.
(1,915 Views)
That module is for reading strain transducers so the input is not specified in V but rather V/V. What exactly are you trying to measure?
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P9yBSAS&l=en-US
(1,911 Views)
Pressure sensor is analog signal 1 to 5 volts.
Customer's older fixture winxp pc, quit working.
Customer ordered NI-9237, so need to find out their sensor type.
If need measure analog input, 0 to 5 volts, what module?
Thanks.
(1,894 Views)
Most of the NI DAQ support +/-10V. You can browse the NI shopping page. On the left, check the C Series Voltage Input Module filter. Then you can select one based on your requirements for channel count, budget, sampling rate etc.
(1,857 Views)
Customer is using the bridge type sensor, mv/V.
It is ok to use it, not my preference.
The original project was on Winxp back in 2002.  Project updated to latest version and pc.
Plot was collecting 300 samples of pressure.  It was 2 seconds for 300 sample points.
Using the library example for Bridge - Continuous Input, it takes ~3000 sample reads for ~2 seconds.
It doesn't matter how low my Sample Rate, Actual Sample Rate doesn't go below 1612.90.
How to get slower scan rate? I need only 300 sample points, not 3000.
Timing Settings
Sample Rate
Actual Sample Rate
Samples per loop
I include my copy of vi.
Thanks for any info.
(1,790 Views)
Accepted by topic author amp61
Hi amp,
You forgot to mention which kind of DAQ device you used with that old WinXP computer!
Now you are using the NI9237 module - and apparently you also forgot to read its spec sheet!
That modules has limits on the sample rate - for both min and max!
The min sample rate  is ~1.6kS/s (for internal sample clock) as clearly written in the datasheet.
To get less samples per second you can decimate the samples. I recommend to set the sample rate to an integer multiple of your desired rate to simplify the decimation…
(1,780 Views)
Thanks!
The original NI board was AT-MIO-16E-10.
Higher scan rate is ok.
(1,764 Views)