NI-9237 analog input channels. Test Panel Input voltage range is -25m to +25m volts.
Can this board allow higher voltages, 5 volts?
If wrong hardware for my application, what DAQ board?
If wrong board, then a sensor in mV range for reading pressure.
Solved! Go to Solution.
07-15-2025 03:15 PM
That module is for reading strain transducers so the input is not specified in V but rather V/V. What exactly are you trying to measure?
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P9yBSAS&l=en-US
07-15-2025 03:41 PM
Pressure sensor is analog signal 1 to 5 volts.
Customer's older fixture winxp pc, quit working.
Customer ordered NI-9237, so need to find out their sensor type.
If need measure analog input, 0 to 5 volts, what module?
Thanks.
07-15-2025 08:04 PM
Most of the NI DAQ support +/-10V. You can browse the NI shopping page. On the left, check the C Series Voltage Input Module filter. Then you can select one based on your requirements for channel count, budget, sampling rate etc.
07-23-2025 04:21 PM
Customer is using the bridge type sensor, mv/V.
It is ok to use it, not my preference.
The original project was on Winxp back in 2002. Project updated to latest version and pc.
Plot was collecting 300 samples of pressure. It was 2 seconds for 300 sample points.
Using the library example for Bridge - Continuous Input, it takes ~3000 sample reads for ~2 seconds.
It doesn't matter how low my Sample Rate, Actual Sample Rate doesn't go below 1612.90.
How to get slower scan rate? I need only 300 sample points, not 3000.
Timing Settings
Sample Rate
Actual Sample Rate
Samples per loop
I include my copy of vi.
Thanks for any info.
07-24-2025 12:52 AM - edited 07-24-2025 12:52 AM
Hi amp,
@amp61 wrote:
Plot was collecting 300 samples of pressure. It was 2 seconds for 300 sample points.
Using the library example for Bridge - Continuous Input, it takes ~3000 sample reads for ~2 seconds.
It doesn't matter how low my Sample Rate, Actual Sample Rate doesn't go below 1612.90.
How to get slower scan rate? I need only 300 sample points, not 3000.
You forgot to mention which kind of DAQ device you used with that old WinXP computer!
Now you are using the NI9237 module - and apparently you also forgot to read its spec sheet!
That modules has limits on the sample rate - for both min and max!
The min sample rate is ~1.6kS/s (for internal sample clock) as clearly written in the datasheet.
To get less samples per second you can decimate the samples. I recommend to set the sample rate to an integer multiple of your desired rate to simplify the decimation…
07-24-2025 03:22 PM
Thanks!
The original NI board was AT-MIO-16E-10.
Higher scan rate is ok.