Hello dmerrtz,
Issue #1:
With regards to the error -200152, this error is often related to
issues with the EEPROM on the device. Whenever we use the DAQmx driver
with the 447x devices, it will always verify the checksum before
performing any task. If the checksum is incorrect, it will throw and
error. You can try to fix this by writting to the "User-defined Info"
property of the EEPROM, which will force the checksum to update. Drop
down a DAQmx calibration property nodes, select "User Defined Info",
now write something and run the VI to execute the task. After you have
done this, reset the DAQmx device.
Refer to this KnowledgeBase:
Issue #2:
Try running the code with highlight execution. Also if you click on the
"Retain Wire Values" and use the probe pointer (get to it by ctr+
shift+alt+ right click, to open the point palette) to select certain
wires to observe the values the get. Make sure you switch you pointer
back.
Regards,
Sandra T.
Applications Engineer | National Instruments