Hello,
Thank for your answer.
In fact, the acquisition in "Slow" mode is not too slow, it just match the spec (p 2-4 of the NI 435x User Manual), witch is given with no auto-zero : 2.8 readings per second, to be divided by the number of channels.
With 5 chanel (+ 1 for CJC), the acquisition take approximately 2.15s in slow mode, no autozero => 2.8 readings/s
Activating Autozero per scan add 1 channel measurement (around 2.5 s => 2.8 readings/s).
Activating Autozero at start should impact only the first scan... but here, it take 0,7 for the whole scan => 8.6 readings/s. That's near from the Fast mode spec : 8.8 readings/s.
My config is a PXI-8106 RT controler with a PXI-4351. I'm Using Labview 8.2.1 with corresponding RT module. And the driver for the 4351 is ni435x 2.5.2 / traditional DAQ 7.4.4
Regards,
Ludos