Hello,
The PXIe-6363 defines -10 to 10V maximum inputs on the input AI channels. Going above these voltages will rail the ADC output to 10V. I think the spec says the absolute max input is 11V.
The question I have is whether this absolute max comes into play only when the CH is activated and an ADC measurement is taken, e.g., in a voltage measurement task? If I connect a given AI channel to a higher voltage, e.g. 16V, but I do not actually start the task or obtain a sample on this AI channel, will this still possibly damage the channel? Or, when not sampling, is the AI in high impedance and the input is protected?
Thanks.
 BertMcMahan
		
			BertMcMahan
		
		
		
		
		
		
		
		
	
			03-12-2024 10:50 AM
The specs here: https://www.ni.com/docs/en-US/bundle/pxie-6363-specs/page/specs.html
say this for overvoltage protection:
I would assume this means that it doesn't matter if you're taking a sample or not, it's got up to +/-25V for two pins while on, and +/-15V while off.
Either way, the AI channels are always high impedance (>10 GΩ) regardless of whether or not it's sampling.
 NIquist
		
			NIquist
		
		
		
		
		
		
		
		
	
			03-12-2024 11:05 AM - edited 03-12-2024 11:07 AM
I suspect it is the opposite, and the card is more vulnerable to damage when NOT operating. Note the specs for overvoltage protection and impedance:
I would avoid connecting larger than +/-11 VDC across any of the pins unless you get a clear OK from NI Support that you can safely do so.