I will receive soon a PZ900 from RKC Japan. It is a temperature controller similar to Eurotherm models.
I downloaded the manuals, drivers, softwares. But the drivers are not for Labview, but for RS422 communication from a software and the PZ900.
I would like to develop my own labview driver for controlling this instrument, and basically, I just need SET T and READ T commands. That's all I need.
Can someone explain me what should I learn to make this driver ?
I will appreciate any advices. I already search, get manuals, etc... I do not request to get a driver already made, I just want to learn the good way to tackle this new problem.
(I have Labview Core 2, I developed a labview program that is already controlling 12 instruments, but this PZ900 is totally different than DMM, lock-in, SMU, etc... so I need to go one more step beyond with labview learning, specifically to this type of instrument PZ900. I guess it is a question of MODBUS or something like this. I am not familiar with that, I probably need a few example to understand what kind of work I should do with the PZ900 to create my driver)
Message 1 of 19
(4,501 Views)
Hi arienai,
Unfortunately you forgot to attach the manual to your message…
In which way is this PZ900 "totally different" to a DMM? For both devices you send a command and then you receive an answer/response!
So you need to open a communication using VISA, send commands and receive the answers. Upon exiting your program you close the VISA reference. You can use any "DMM"-type device driver for inspiration, like those for Keithley2000 devices…
(How does MODBUS come into play?)
Message 2 of 19
(4,464 Views)
Dear GerdW,
here are the links :
https://www.rkcinst.co.jp/english/downlaods/8918/imr03b07e1/
https://www.rkcinst.co.jp/english/wp-content/uploads/sites/2/2019/06/imr03b05e1.pdf
This instrument will not be on localhost with TCP. It will be plugged to the computer by USB connection. (through the maker module COM-K1 , to convert RS422 to USB).
Today, I downloaded and installed Modbus library on my Labview 2019 (Win 10, 64 bit). I found an example based on TCP. I need to convert this example to SERIAL.
In the manual, section 6.4 of the document imr03b06e1.pdf, it deals with single word data. I found item 1 related to "measured value (PV)", related to "Register address 0", and also "Set Value (SV) monitor", related to "Register address 1".
Looks quite simple. I will try during this week. The machine is arriving tomorrow.
Yours
Message 3 of 19
(4,436 Views)
Hi arienai,
Which Modbus library did you download and install? Most often they provide examples for serial communication (Modbus/RTU) as well…
Did you configure your device to use Modbus communication as described in the manual?
You did not provide a link to "imr03b06e1" so I can't tell about setting SV or reading PV data…
Message 4 of 19
(4,426 Views)
Sorry, I though it could be reached through the links I mentioned.
I attached those documents in a ZIP file.
So far this is what I have, but I am not so familiar with the content ...
I will see when the PZ900 will arrive in the lab tomorrow .
I will connect it to a desktop through USB.
Once again, thank you for taking time to help me, or simply to read those docs, cause the content is a little bit complicated. But maybe it is not that difficult by using modbus toolbox.
Message 5 of 19
(4,421 Views)
I received COM-K2 module to plug the PZ900 (Modulus/RTU) to a PC via USB cable.
The PZ900 can be controller by an available software PRO-TEM from RKC. To use it, I have installed the drivers of the com module COM-K2, then run PRO-TEM, and I could send commands to set temperature, and read temperature. However, I have no idea about the structure of the instruction.
The attached programming guide, is helpful to understand.
The PZ900 is detected as COM10 on device manager (Win10 - 64bit).
However, I could not send command with VISA like I used to do for DMM or other instruments that are "Labview compatible" (with LV  drivers ).
So I guess I have to send hexadecimal instructions to be able to communicate with this PZ900.
I will try to do it. But if you have a small example of labview vi that could do that, I will be very grateful.
Message 6 of 19
(4,364 Views)
Hi arienai,
your manual describes the "PLC communication". You should read the "host communication" manual instead, as mentioned on page 1-2. (It mentions "Modbus"…)
Message 7 of 19
(4,362 Views)
Thank you GerdW, I attached the host com manual.
I have to read it, but if I understand the principle of creating 1 instruction, then I can (maybe?) do similarly for all commands and create a driver for PZ900.
A short example of write/read/query will be more than appreciated. I hope I will be able to send 1 command to PZ900 using this Modbus protocol.
Example 1 : set temperature  239.5 (degree C), infinite time duration
Example 2: ramp temperature from initial Ti to final Tf in 10min
Example 3 : read temperature like 573.2 (deg C) after conversion from hexadecimal to decimal and the decimal point is not recognized...
This is almost what I need for the moment. Next, I will have to define PID ... which is maybe more complicated.
Yours
Message 8 of 19
(4,351 Views)
Hi arienai,
your device supports two different type of protocols: the "RKC" one and a standard Modbus protocol.
You may implement the RKC-type protocol as described in chapter 4, but I suggest to use Modbus with one of the available libraries. All you need to learn is to use the correct Modbus registers and the correct data encoding - as described in chapter 5 & 6…
Message 9 of 19
(4,342 Views)
Dear GerdW
MODBUS order of data transfer is set from "high-order word to low-order word".
Actually, the settings are for input data type = DOUBLE WORD.
The number of measured value digits is 5.
(The com data type can be checked in Modbus double word by : 013EH, 013FH )
Could you just tell me the difference between programming with "high-order to low order" and "low order to high order" ?
What is the difference between programming in "double word" and "single word" ?
The master is the PC.
The slave address (PZ900) is 01
The parameter ADD (device address) has been set to 1 in order the master to communicate with the slave.
I noticed the slave address is 1, but it is 2bits, so it should become 01.
In MODBUS, the message format is :
slave address : 01
function code : 03H (to get temperature )
data
error check (CRC-16)
Based on page 6-3-1, double word, could you confirm that the correct sequence of instruction should be  :
- measure value (PV) : 01 03H 00H 00H 00H 04H 44H 3AH  --> is it correct ?
- set value (SV) monitor : ??
Actually, I am interested in FIX set mode control : set temperature T.
But also by a RAMP in temperature until Temp T, followed by a plateau at T with infinite duration (until setting another temperature).
The reason of the RAMP is that I can control the amount of power to prevent damages on the hot plate (hardware)
As mentioned in 5-13, there is a warning for the decimal numbers, as MODBUS does not recognize data with decimal point.
Then, I cannot understand how PZ900 can understand a set value 298.5 for example. There is 5 digits and 4 bytes data, so it is OK, but how PZ900 can interpret the decimal part ?
So here is back to orignal question about the difference between programming in single word rather than double word : is single word prevent decimal settings ?
Further, ...  the MODBUS data mapping will break my brain ... for sure !
Message 10 of 19