RTD measurement
I'd like to measure the temperature with the use of PT100, using RTD01.
Anyone can tell me how configure the DAQ channel?
I've labview 7.1.
Thank-you very much
Roby
(2,833 Views)
Hi Roby,
To configure a channel for your RTD01, you will first have to set up your SCC carrier.
You can do this in MAX (Measurement and Automation Explorer) by right-clicking on "NI-DAQmx Devices" and creating a new "NI-DAQmx SCC Connector Block"
Then modify the properties of your SC-2345/2350 to add your SCC-RTD01 to the appropriate location.
Now set up a DAQmx task for your temperature input:
1) MAX -> Data Neighborhood -> NI-DAQmx Tasks -> Create New -> Analog Input
2) Choose either Temperature -> RTD or Voltage (you can set up a custom scaling to convert voltage to temperature later)
3) Select the appropriate physical channel on your RTD01
4) Enter a name for this task
5) Configure the RTD/Voltage settings accordingly
You've now conf
igured a DAQ channel!
You can find detailed help on RTD measurements at the following ni.com link:
http://zone.ni.com/devzone/conceptd.nsf/webmain/D9BAA2AA872A6B5F86256D730079E3E6?opendocument
Good luck with your temperature measurements,
Andrew A
(2,833 Views)