SCC-AI03, can I use this module to measure current?
I have a couple SCC-AI03 module that read voltages but I have some pressure transducers that have current readings, will it be possible to use the SCC-AI03 to read current values?
Thank you,
Message 1 of 8
(3,993 Views)
I'd assume you could just use a shunt resistor. Check the how to guide: https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/current-measurements-how-to-gui...
Message 2 of 8
(3,989 Views)
Hello mike-o-tronic,
I have a module that reads 4-20 mA with an output of 0-10V, do you know what kind of shunt do i need to make this work?
Thank you again,
Message 3 of 8
(3,982 Views)
My bad! I thought you had a current output and a voltage input. I'm not sure if you could measure voltage with a current input. I'm also not sure if the SCC-AI03 would work with a current input. Not that I have any experience with that module but judging from the manual.
Message 4 of 8
(3,961 Views)
That's because he said in the first message he had AI_03 module which read voltage (it does +/- 10V) with a transducer that has current "readings" (which doesn't make sense as a phrase.  A transducer OUTPUTS a signal.
Now the last message is confusing because it says "I have a module that reads 4-20 mA with an output of 0-10V".  An SCC module can be an analog input or output (and be voltage or current) but it can't input AND output.    So if he has a module that reads current, then it can't be an SCC-AI03.
Clarification is needed.
Message 5 of 8
(3,952 Views)
Indeed clarification is needed. Sorry for all the confusion. Here is the situation:
I have a pressure transducer that outputs 4-20 mA, then I have some modules models SCC-AI03 that reads analog inputs in voltage (0 to 10V) and I was wondering if I can use the SCC-AI03 to read my pressure transducer that outputs a signal in Amps.
Thank you,
Message 6 of 8
(3,940 Views)
Accepted by OJ1990
Yes.  Put a 500 ohm resistor across AI+ and AI-.  That allows you to measure the voltage drop across that resistor..  For 20 mA, voltage drop is 10V.  For 4 mA, the drop is 2 V.  Now you've converted your 4-20 mA signal to a 2-10V signal.   Basic electronics.
Message 7 of 8
(3,935 Views)
Thank you all for the help!!! Greatly appreciate it.
Message 8 of 8
(3,929 Views)