03-19-2019 08:23 AM
Hi,
A piece of Keysight E5071C codes was resued for E5080A with code Emulator.
There were 3 traces on the screen at the beginning. Command
:CALCulate:PARameter:COUNt 5
is working with Keysight I/O but not with NI Max. ENA5080 has error message SCPI 106 with NI Max same as Labview.
Keysight E5080A FW A12.60.03.
Any suggestion?
Thanks,
GTR
03-21-2019 01:37 PM
Hello GTR,
When you write *IND?\n in a NI-VISA Test Panel, what reads out? Can you show us a screenshot of your error message? A little more information will be helpful in narrowing down the issue.
03-21-2019 03:42 PM - edited 03-21-2019 03:44 PM
You didn't say what interface you were using, but I have found that Agilent/Keysight needs to have the "Tulip Passport" enabled in NI-MAX.
This is the only "Passport" that is disabled by default
Open NI MAX
Navigate to Tools---->NI VISA---->VISA Options
Select Passports from the tree
Check the box to enable NiVisaTulip.dll -- NI-VISA Passport for Tulip