Thank you for getting back to me so fast. I was wrong about the series, the 6251 is a M. I don't understand how I am getting bleed through. The card supports 1Gs/s across multiple channels. So if I am measuring two channels I should be able to measure at 500 ks/s per channel, correct? The sample rate I am using is only 10 k/s. I tested it, as I change the sample rate the values do change. How can I get around this?
Thanks
08-22-2005 01:04 PM
05-28-2006 08:01 AM
05-30-2006
11:24 PM
- last edited on
08-06-2026
10:31 AM
by
Content Cleaner
Steini,
Have you attempted to change the convert clock rate as mentioned in the above KnowledgeBase? A slower convert clock rate will increase the time between samples, giving the instrumentation amplifier on your DAQ card longer to settle. Even at low sampling rates, the convert clock rate is still quite fast. Depending on the magnitude of your "interference," this may or may not be the cause. Another possible cause could be the specified voltage range of the input channel. Are both channels set for the same input ranges? Also, what happens when you read a grounded channel in between your first and second channels (Connect ai2 to ground and set your scan list to ai0, ai2, ai1)? Does this fix the second channel? Is the first channel ever inaccurate? Please let me know the answers to these questions, and I would be more than happy to help you continue troubleshooting.
Hope this helps,
Message Edited by Voltage Viper [DE] on 05-30-2006 11:24 PM
05-31-2006 07:21 AM