Hi,
Check that default instrument settings are the same as your top level VI expects, particular if you are not using default settings. Cal Houses tend to use Instrument default addresses and then dont return the addresses back to original settings
Do you trap on the error cluster from the Instrument Initialisation VIs. May be use some breakpoints or open up some lower level VI's to see what's going on.
How are you communicating with you instruments, RS232, GPIB, DAQ?
Regards
Ray Farmer
08-09-2005 07:04 AM
What is your configuration? You say that you have a card in the PC connected to another card, you mention instruments, etc. For a card in the PC, if it is a National Instruments card you can use the Measurement & Automation explorer tool that installs with the DAQ drivers. As mentioned regarding external instruments, if they are addressable be certain that the addresses are correct. You say it worked until last week but you haven't used it for a month which is confusing. What is the time frame for the calibration vs when you detected the problem? Did it work before cal, after cal, a while after cal? What instruments, how are they connected, etc.?
P.M.
08-09-2005 08:43 AM
Before we get too in depth I must explain that I am a novice with this software and didn't set it up I've have inherited it.
We are communicating using DAQ, I believe it's a PCI-M10-16E-4 I/O Board. The Pressure (IOMEGA PX223) and Temperature (5B37 Thermocouple) Instruments are wired to a carrier which is connected to the PCI Board in the PC.
The last time the equipment was used until this problem on Monday was about one month ago, everything worked ok. Previously to last month the program has been working on and off for four years.
Where do I find the Instrument addresses?
Rob