Time out error using inst.llb vi (Agilent 2000 3000 X-Series Acquire Waveform Continuously) with a Agilent MSO-X 3054A
We are having an issue with a vi in the instrument library. It is (Agilent 2000 3000 X-Series Acquire Waveform Continuously). The problem is it works ok with our DSO-X 2002A scope with no issues. It don’t work with our MSO-X 3054A scope. See error below. Sometimes it works but most often it doesn’t and it locks up NI-VISA when it does fail. I’m connected to the instrument using USB.
Error -1073807339 occurred at VISA Read in Agilent 2000 3000 X-Series.lvlib:Wait for Acquisition Complete.vi->Agilent 2000 3000 X-Series.lvlib:Read Waveforms (Single).vi->Agilent 2000 3000 X-Series Acquire Waveform Continuously.vi
Possible reason(s):
VISA: (Hex 0xBFFF0015) Timeout expired before operation completed.
(3,786 Views)
Here is a KnowledgeBase article with several solultions listed for this error. Give these a look and see if any of these solutions look like they will solve the problem you are experiencing.
http://digital.ni.com/public.nsf/allkb/874B379E24C0A0D686256FCF007A6EA0
Best,
Doug
Aaron Douglass
(3,746 Views)
Thanks for the idea. Unfortunately I had no luck. I’m running NI-VISA 15.5. I added 10 seconds to my timeout and it still fails. The pit falls of using VI’s from the inst.lib sometime it works other time you spend more time troubleshooting.
(3,727 Views)
Accepted by topic author calmeyn
I solved the problem. The scope was running an old version of software 2.20 curent is 2.4X.
(3,668 Views)