I currently am running windows 11 on my laptop. I have the 2022 Keysight connection expert downloaded and can communicate to my VNA via the connection expert. However, when I open my NI-MAX window, I cannot see the VNA as one of the possible options to connect to. I have downloaded all the drivers such as NI-VISA, NI-488.2, ect. I even noticed on some forums that I may have been required to tick the "enable keysight GPIB Cards for 488 programs" and I even ticked the box in the NI-MAX settings where I had to tick the box to enable tulip. After all this, I am still unable to make LabVIEW pickup/connect to my VNA P9375A device. Please let me know if there are any additional steps that I may have missed or may need to still do.
 Teekayy
		
			Teekayy
		
		
		
		
		
		
		
		
	
			09-12-2023 05:36 PM
Hi OSHAMOUT,
Were you able to figure out your problem because I ran into a similar problem initializing the VNA P9375A?
 rolfk
		
			rolfk
		
		
		 
		
		
		
		
		
	
			09-13-2023 02:15 AM
Have you even connected your VNA through a GPIB card? Typically those devices are nowadays connected through Ethernet or USB. USB seems the only option for this device. The "enable Keysight GPIB cards" option (and the Tulip driver) only applies to when you use a physical HP/Agilent/Keysight GPIB interface card in your PC.
You need to have the Keysight IO Suite driver software installed: http://www.keysight.com/find/iosuite
 Teekayy
		
			Teekayy
		
		
		
		
		
		
		
		
	
			09-13-2023 09:55 AM
Hi Rolf,
I'm connecting to the VNA through USB-C. Keysight connection expert confirms the connection but when attempting to initialize the device through Keysight NA labview drivers it throws a I/O error. Perhaps my resource string is incorrect.. PXI0::0-0.0::INSTR
I also have installed Keysight IO suite, IVI .net and C drivers along with labview drivers for the VNA. Would you have any suggestions or solutions?
Thank you,
TK
 rolfk
		
			rolfk
		
		
		 
		
		
		
		
		
	
			09-13-2023 10:12 AM
That resource string definitely looks wrong. Neither is it a PXI chassis nor does an address 0.0.0 mean anything.
 Teekayy
		
			Teekayy
		
		
		
		
		
		
		
		
	
			09-13-2023 11:58 AM
Hey Rolf,
I've solved it. Apparently the previous visa address was for the instrument itself but the one I needed was a hiSLIP so I needed to add TCPIP0:: and my connection::hislip_PXI10_CHASSIS1_SLOT1_INDEX0::INSTR. for remote interface. It runs now 🙂
Thanks,