Vi for fast tc data acquisition
Does someome have an example vi for fast data thermocouple acquisition using a ni9212 device?
(8,022 Views)
What is your definition of "fast"? How is this module connected to the rest of your system?
(8,001 Views)
The ni 9212 module is inserted in a cDAQ 9171 SLOT which is usb  connected to the pc. For "fast", I mean a sampling rate of 50 points per second. The examples I've found in the labview libraries are not so fast.
(7,974 Views)
Accepted by Winzy68
You need to set ADCTimingMode to High Speed. See Can I Sample at More Than 2 Hz With My NI 9212, 9213, 9214, and 9219?
https://tmesystems.net/
(7,958 Views)
Why?  There are VERY FEW cases that result in a delta temp measurable @50hz!  You really need to understand the theta jc (coefficient of thermal conductivity) of the unit under test.  If theta jc really is that high, it is likely that the mass of the thermocouple itself will swamp the system measurement because its mass will slow down any change in the UUT's temperature.
You neither want nor need high speed tc measurements. 😀
(7,918 Views)
The other case where the temperature rise is that fast would result from an explosion that probably obliterated the thermocouple.  In that case again, speeding up the measurement is not your biggest problem.
(7,913 Views)
Thank you very much! Now the system works properly!
(7,836 Views)
