No. It looks like you may have the 40.5 and the 1,000,000 inverted. While you add 25 (I assume you are looking at deg C?), you don't handle the note that says 1 mV output at 25 deg C on page 2 of the manual.
Do you have the reference bath set up?
If the output from the read VI is in Volts, subtract .001 (for 1 mV). Divide by 1 million. Multiply by 40.5. Add 25.
Yours, you take 40.5 and divide by 1 million, then wind up dividing that into the output of your read function.