12-18-2009 04:57 AM
12-18-2009 05:02 AM
First of all this would fit better in the hardware forums.
- What software are you using, so where do you want to change this settings?
- Do you want to change the input range (+/- 10V) or the terminal configuration (RSE, NRSE, Diff), or both?
Christian
12-18-2009 05:40 AM
12-18-2009 07:23 AM
"...sometimes the 6008 starts up in a mode where the input range is less than 5V"
The device is just doing what you tell it should do!
You can change the min/max values as well as the terminal configuration at the "Create Task.vi"
If you are using the Express VI DAQmx Assistent you set this in the task configuration.
Christian
12-18-2009 07:31 AM
I am using "DAQmx Create Virtual Channel " to set differential mode and "analog voltage input". I leave minimum and maximum value "open" using the default 5V value. Should'nt that be okay?
Ole
12-18-2009 09:49 AM
12-20-2009 05:51 AM
How do you determin that the device starts with another range then specified?
You could use the channel property node to read the actuall configuration btw.
Christian
12-22-2009 02:17 AM