how to avoid floating values rather than zero when atempt by a pulse

Hello,

 

I'm using the 9401 with a module 9223 to reading and measuring time and amplitude based signals. My problem is: The value of sensors without any interference presents variation from -0,003 V up to +0,004 V and these variations aren't constant. This mean that may be -0,0008; -0,0015 for negative values and +0,0022; +0,0018; +0,0031 for positive values. So I can't read the Zero value (+/- 0,0075 mV) that could be acceptable tolerance. I'm using a piezoelectric sensor for acoustic emission measurement without filter and preamp connected. it has a response frequency of 20 kHz up to 600 kHz and maximum amplitude of -65 dB. Thanks for any help.

B. Regards 

0 Kudos
Message 1 of 8
(3,081 Views)

With noise you'll usually need to do averaging or filtering to get decent measurements. How about the average of a 10 slot queue of the latest measurements? That or the simpler alpha-filter (x% of the latest measurement + 1-x% of the previous value, 20-40% is often a good starting point)

 

/Y

G# - Award winning reference based OOP for LV, for free! - Qestit VIPM GitHub

Qestit Systems
Certified-LabVIEW-Developer
0 Kudos
Message 2 of 8
(3,067 Views)

I didn't understand very well your solution because I'm newer in all NI solutions. Generally I use a VISHAY data system rather than NI. If I'm right, you mean that a filter can adjusts it. Actually the sensor is plugged in a isolated surface without noise, so the theoretical amplitude value might be Zero or +/-0,0001mV, but the soft receives (in self test mode) a floatation wave and may be due to impedance or ground interference. I don't know. I tried to ground the cable but without good result neither with a 1000 OHM capacitor. The signal is similar to a oscilloscope without calibration. May the NI-9223 isn't calibrated too??

0 Kudos
Message 3 of 8
(3,053 Views)

Why your measurement is floating i really cant answer. With such small measurements it could be antenna effects in cables or a number of different reasons. A way to reduce the effects is to average a number of samples. Many multi meters use this approach (or has the ability).

/Y

G# - Award winning reference based OOP for LV, for free! - Qestit VIPM GitHub

Qestit Systems
Certified-LabVIEW-Developer
0 Kudos
Message 4 of 8
(3,047 Views)

Hi!

 

NI 9223 is a 16bit device and its maximum accuracy is 0.003V. If you read fluctuations below 0.003V it is a 100% expected behavior.
If you need more accuracy, you have to choose a 24bit device. 

MarcoMar

National Instruments Italy
0 Kudos
Message 5 of 8
(3,031 Views)

@Ortenzi wrote:

Hello,

 

I'm using the 9401 with a module 9223 to reading and measuring time and amplitude based signals. My problem is: The value of sensors without any interference presents variation from -0,003 V up to +0,004 V and these variations aren't constant. This mean that may be -0,0008; -0,0015 for negative values and +0,0022; +0,0018; +0,0031 for positive values. So I can't read the Zero value (+/- 0,0075 mV) that could be acceptable tolerance. I'm using a piezoelectric sensor for acoustic emission measurement without filter and preamp connected. it has a response frequency of 20 kHz up to 600 kHz and maximum amplitude of -65 dB. Thanks for any help.

B. Regards 


Then dealing with ADC equipment. The least significant bit(s) will always flicker some. So you will never read a stable zero value. Even if you tie a input channel to ground and do some measuring. The next thing is the ADC resolution. For your device that is (20 volt)/(2^16) equal to about 0.3 mV. So your " can't read the Zero value (+/- 0,0075 mV) that could be acceptable tolerance" Do not make any sense in your setting. Then it is the term Absolute Accuracy you can read more about it here http://www.ni.com/advisor/accuracy/acc_support.htm for your device this number is 3mV. Another thing to point out. Is that you say. You do not use any preamp. I guess the output from sensor is quite low. So you should use an amplifier. And hence make use of full ADC range. To conclude I would say you have some good tools in your toolbox. But since you are not using them correct. Your output would be more or less gobbledygook 



Besides which, my opinion is that Express VIs Carthage must be destroyed deleted
(Sorry no Labview "brag list" so far)
0 Kudos
Message 6 of 8
(3,017 Views)

Thanks for all. It is a good lesson to learn. You must only buy a product that you can proof it's good for you.  

0 Kudos
Message 7 of 8
(3,012 Views)

Si Marco, lo so. I think I sent you the sensor to test with your instruments in Italy prior to return to Brazil. Actually, I couldn't buy this module because it isn't good for my purposes. By anyway I'll resell it here and I'll buy another kind of hardware. tks.

0 Kudos
Message 8 of 8