@nehehdjet wrote:

Hello,

 

thanks a lot for the vi linked by shredderofmass.

 

I got it working with our AQ6317 and Labview 2010. Some hints:

 

Before you are able to start the program, your AQ6317 needs to have a VISA adress and needs to be able to communicate with your pc. To check this, use the NI Automation and Measurement (NI Max). You can change the VISA adress of the AQ6317 by pushing the "System" button and then MY ADR.

If your Ando is not found by the NI MAX, switch it off and on again. If you plug in the connection to the computer while the instrument is running, it does not seem to work.

 

The VI uses points as a decimal delimiter. If you are in europe and your pc uses kommas, nothing will work! You can change this in your Labview in "Options".

 

Thanks a lot to you all!

 

Best regards, Tina


Thank you for your input, Tina!  Years later, your advice helps me.  Cheers. 