 kwanghyun92
			kwanghyun92
			02-07-2017 07:07 AM
Hello.
I am trying to measure frequency sweep data using function generator(agilent 33250A) and oscilloscope(DSO-X-4024A).
But I have some difficulties making frequency sweep when I use "configure frequency sweep.vi".
The difficulty is that frequency sweep is repeated automatically until I will stop the program.
Before describing my problem, Let me explain "configure frequency sweep.vi".
I should give some data to "configure frequency sweep.vi", 'start frequency', 'stop frequency', 'sweep time', to use this vi.
For example, If I start the "configure frequency sweep.vi" with 3700Hz(start frequency), 3800Hz(stop frequency), 100s(sweep time), the program will be repeated automatically after the frequency reach the stop frequency.
Let me describing my problem.
I want to make the frequency signal will be held when the frequency sweep reach a stop frequency that I set.
For example, If the frequency reaches the stop frequency(3800Hz), the function generator will emit the signals(3800Hz) until I stop.
I want to discuss this problem with you.
I attach a block diagram of "configure frequency sweep.vi".
 jamiva
			jamiva
			02-07-2017 07:26 AM
This is from the manual. It looks like you need an external trigger to do what you want.
02-07-2017 07:43 AM
Thanks for your replying.
I knew the fact.
but the external trigger provides only 1 time frequency sweep and output will be not held after 1 time frequency sweep.
 jamiva
			jamiva
			02-07-2017 08:25 AM
I don't have an Agilent 33250A. Quickly looking at the manual, I think you have a hardware limitation with that SigGen model.
I do have a slightly newer model (Agilent 33521A) and it has two additional sweep parameters: "hold time" and "return time". I was able to easily create your waveform. Probably time to either redefine your test waveform, or find other hardware that meets your needs.
 jamiva
			jamiva
			02-07-2017 08:43 AM
I was thinking that you may have another option:
It looks like you want to scan from 3700Hz to 3800Hz over a 100 sec sweep time. That's a 100 Hz delta over 100 seconds, or a sweep rate of 1 Hz/sec
You could programmatically create a loop to update the frequency by 1 Hz every 1 sec. That is if you can live with the discontinuities of a stepped sine wave. Also, the hardware will not switch frequencies at exactly 1 sec intervals. It take a finite amount of time to send the command, process the command, and for the waveform to settle. However, the advantage of doing it this way is that you could synchronize/trigger the o'scope to capture a CW waveform during that 1 second interval.
02-07-2017 09:03 AM
After reading your replying, I check the manual of Agilent 33250A.
Unfortunately, this model doesn't have the function("hold time" and return time".)
02-07-2017 09:07 AM
Actually, I made the program as you said above(prgrammatically repeat a loop to update the frequency by 1 Hz every 1 sec.
But, I want to make the frequency increase with continuously data.
I mean that there will be so many data between 3700Hz and 3701Hz.
(3700Hz, 3700.0001Hz....3700.9999Hz)
But the program that create a loop to update the frequency creates only point of signal.
(3700Hz, 3701Hz)
So I am trying to make a program using "configure frequency sweep.vi".