The attached vi runs without error with the PXI-5102 Digitizer. When run with the PXI-5114, I get niScope Configuration Clock.vi Error 1074118538. The error description is "Specified source or destination is invalid". I am using LabVIEW 8.0 and niScope 3.0. Any ideas?
Thanks in advance.
07-21-2010 01:04 PM
Hi VI Joe,
Some of the VIs you have in there are 5102, specific. Best way to do synchronization with the 5114 is through TClk. A good example can be found in the LabVIEW Example Finder under Hardware Input and Output->Modular Instruments->Digitizers->Synchronization.