04-19-2017 09:16 PM
I already try for your version.but currently, still same result..only the ch.2 didnt get the reading..I try figure out with different method..there any chance the problem from labview or the hardware it self?
04-19-2017 09:31 PM - edited 04-19-2017 09:51 PM
it shows error -1073807339..source VISA read in readvalues.vi. regarding to SW2.I'm using the RS232. daq NI USB 6341
04-19-2017 10:00 PM
That error is due to a timeout.
I'm a little confused - it did the same as yours (no result for Ch2, but everything else fine) or it produced error -1073807339? Or both?
Regarding SW2 - there should be a series of switches on the hardware (the amplifier, not anything related to NI). Can you tell me their positions? For example, 1 up, 2 down, 3 down, 4 up... etc
04-19-2017 10:33 PM
both. when first time I try, its shows zero reading for ch.2. for the 2nd time I try its show the error..here I'm attach the configuration switches for your reference.and the SR530 manual.
04-20-2017 12:55 AM
04-21-2017 01:03 PM
Here the switches on my amplifier that I choose.
04-23-2017 05:55 AM
We're getting closer, but as it says in the manual that you linked earlier (emphasis mine):
There are two banks of 8 switches, SW1 and SW2, located on the rear panel. SW1 sets the GPIB address and SW2 sets the RS232 parameters.
Perhaps a photograph of the back (or just a description) of those switches (SW2 - we don't care about SW1).