07-02-2012 11:27 AM
Hello, i have a "NI BNC-2090 - Rack-Mounted BNC Terminal Block" + "Labview 10" + "NI 6289", i'm trying to output a digital trigger, but in vain, the DAQ assistant inform me that the output trigger is on the PFI12, but the BNC 2090 juste have PFI1 to PFI9, so my question is, is it possible to modify the output PFI12 to PFI9 for example ?
Thank you
Cordially
Ps: I tried all examples refer to the digital triggering in vain.
07-04-2012 01:33 AM
HI,
Thanks for posting on NI forum.
If you look the 6289 pinout you can see the PFI12 is CTR0 Out :
Try to connect the output "CTR0OUT" of your BNC 2090.
Best Regards,
Nicolas M.
National Instruments France