take readings from Lakeshore 336 T controller through VISA TCPIP, but very slow

I connected a Lakeshore 336 Temperature controller to a PC ethernet card by using a normal RJ45 cable.
The VISA resource name is set as "tcpip::192.168.0.12::7777::socket". But the readings is very slow. It takes about 2 seconds for reading one sensor data. But the manual says "10 readings/second".

I tried both Lakeshore 336 Temperature controller and PC connected to a router, but the same slow speed.
How can I get reading as fast as GPIB interface using Ethernet port?

 

Thanks.

LV2013SP1@Win7Pro
WT

0 Kudos
Message 1 of 3
(4,222 Views)

Attached are the snapshots of the front panel and block diagram (with 2 probes) of the VI reading LS336 sensor.

There are 2 probes inserted. One is right after VISA Write, which shows no error; the other one is right after VISA Read, which shows error. But strangely,  the VI still acquires the sensor readings correctly even the error occurs.

Thanks in advance.

 

WT

 

front panel.PNG

block diagram.PNG

0 Kudos
Message 2 of 3
(4,172 Views)

By several days trials, I found the reason.

The byte count set to the VISA read function have be to equal or less than the return bytes by the instruments. Otherwise, timeout (slow response) happens.

I don't understand why VISA-TCP/IP cannot work like VISA-GPIB which can automatic accept the returned bytes no matter how many byte count I set to the VISA read function.