tds3000 signal check

I am running LabVIEW 8.2.1 and using a Tektronix TDS3034B to read some signals. I have the O-scope connected via GPIB-USB cable and I occasionally lose all communication with the scope. Is there any way to ping the scope to be sure I have connection with it before beginning my test? A lot depends on receiving a signal from the scope! Any help would be great! Thanks!
0 Kudos
Message 1 of 6
(2,671 Views)

The basic SCPI (GPIB protocol) command for this is to send *IDN?

 

That should return the Instrument name string.  If it's returned, you're communicating.

LabVIEW Pro Dev & Measurement Studio Pro (VS Pro) 2019
0 Kudos
Message 2 of 6
(2,666 Views)

Thanks for your response, I guess I should have been a bit more specific though. I'm trying to do it programmatically so that I can check the connectivity each time a button is pressed.  Is there a way to send the identification string programmatically and get a boolean response? 

 

Thanks!

0 Kudos
Message 3 of 6
(2,659 Views)

I have had the same exact problem with a Tektronix scope or function generator; I do not remember which one.

 

Some Tektronix instruments seem to have a problem with USB control through LabVIEW. I would lose contact with the instrument if I was using USB communications, however, same instrument same program with USB to GPIB did not have that problem.

 

I would suggest using the FIND VISA Resources function to see if your instrument is still attached; to make it into a boolean, use the MATCH String Function with -1 being false. (You should be able to find the VISA device ID in MAX.)

 

If your program is critical, then try GPIN if possible to avoid this problem.

 

Cheers,

Andrew

Message 4 of 6
(2,655 Views)

*IDN? can easily be used programatically.  You just send it to the instrument using VISA, GPIB, serial or whatever; and then read the return string.  [see example LABVIEW - GPIB.vi]

As mentioned Find VISA Resource will work too (it uses *IDN? internally).  Or, the GPIB status VIs will also verify communication.  The Instrument Drivers you're using for your TDS 3000 series scope should have a simple VI that works as well but I don't use that model so I'm not sure.  The 5000 and 7000 series drivers have revision check VIs that will return the firmware version, and the initialize VI can be set to return the ID string.  All these techniques will verify your connection.

 

BTW, I've also had trouble with Textronix USB comm but NOT when using a USB-GPIB interface, only plain USB.

LabVIEW Pro Dev & Measurement Studio Pro (VS Pro) 2019
Message 5 of 6
(2,635 Views)

Find VISA Resource does not use *IDN? internally. It reads the visaconf file which is created by MAX.

 

If the USB-GPIB goes off line as described in the first post, any command is just going to return an error. Even with a successful response, there would be no guarantee that the connection would not be lost in the middle of the test. You want to eliminate the lost connection problem. I did have the same problem and fixed it by using a quality USB hub or by connecting directly to one of the PC's USB ports. Also make sure there is no power saving options turned on.

Message 6 of 6