What sort of values do you actually get? How have you configured the acquisition (rse, diff) and what pins did you connect? What do you read in the MAX test panel?
0 Kudos
Message 11 of 20
(2,541 Views)

i got values between -95 and +45.Firstly i asked, what is the difference between diff and rse which one should i do?i connect acquire data-analog input-voltage-ai6

 

0 Kudos
Message 12 of 20
(2,538 Views)

and acquisition mode is continuous sample

0 Kudos
Message 13 of 20
(2,535 Views)
Diff = differential
Rse = referenced single ended

You should be using differential.

I asked for pin numbers, not names but you only provided a single name. No idea of you connected gnd or the second channel of a differential connection.
0 Kudos
Message 14 of 20
(2,515 Views)

pin1:+

pin3:output

pin4:-

 

 

0 Kudos
Message 15 of 20
(2,503 Views)
That does not make any sense. For a differential connection, you connect ai0 and ai8 - pins 1 and 2 with the usb device and the screw terminals. Did you look at the connection diagram given in MAX or the DAQ Assistant?
Message 16 of 20
(2,496 Views)

This my DAQ pinout.terminals.png

 

I connect signal to ai0.for differential connection, where am i suppose to wire power supply - in this DAQ?Could you please comment on this figure.For ex. there exist AI8 but in DAQ surface i cannot see it

0 Kudos
Message 17 of 20
(2,474 Views)

In MAX i tried to look at it but is says; the connection diagram tool does not support this device

0 Kudos
Message 18 of 20
(2,470 Views)

http://www.ni.com/pdf/manuals/375204a.pdf this is my device's manual.at page20-21-22 say something about you said but could you please physically tell me how to do it?

0 Kudos
Message 19 of 20
(2,458 Views)

i have solved it. ai0- is round metal cylinder 🙂

0 Kudos
Message 20 of 20
(2,439 Views)