12-08-2008 02:49 PM
My Agilent 34410A DMM stays in remote mode after my code exits, or if I use MAX to communicate with it, but only if I'm communicating over USB. If I use GPIB, the DMM returns to local mode when my code closes. Apparently there is no LOCAL command for this instrument. Is there any way, other than pressing the Local button, to put the instrument back in local mode using a USB interface? I call hp34401a_close() to close the instrument.
Thanks for any suggestions.
12-09-2008 10:45 AM
Hi JoeN,
I think what is happening is that when using the GPIB interface you are sending a code to set the instrument back to Local. I took a look at the user's manual for the hp34401a and it seems to have a command to set this back: SYSTem:LOCal.
You can try this at the end of your USB interface program, and see if this is able to help you out. Otherwise I would run NI-SPY on the GPIB end and see exactly what command is being sent to disable remote control of the device.
Let me know if this helps
12-09-2008 01:41 PM
Hi Can,
SYST:LOC doesn't work for the 34410A, and it's not listed in the 34410A Programmer's Reference. When I send the command the instrument displays "ERROR", and returns error code -113,"Undefined header". It turns out that SYST:LOC is listed as a command that only works on the 34401A over RS-232, and the 34410A doesn't have an RS-232 interface, so I guess I'm out of luck.
Thanks very much for taking the time to try to answer my question.
Joe
12-10-2008 03:54 PM
03-09-2011 05:59 PM
Hey JoeN
Did you find a solution to this. I am having a similar problem when I used the LAN interface.
Arun
03-10-2011 02:36 PM
Hi Arun,
Did you try out the suggestions posted above?
03-10-2011 03:43 PM
Yes I tried SYSTem:LOCal but I got the same error that Joe N got. I am using the Agilent 34410A. For some reason it just remains in the remote mode and I can't figure out how to get out of with without having to press the local button on the instrument.
Arun
03-11-2011 09:50 AM
Hi somanath,
Have you checked the manual to see if there was a diffrent command? Perhaps the device is not capable of being set this way. and can only be set manually by using the button. It might be a good idea to contact Agilent to see if your device has this capability.
12-09-2011 04:13 PM - edited 12-09-2011 04:16 PM
I found a way to do it.
If you use the VISA GPIB Control REN VI and set the input to "Go To Local".
It is found in Instrument IO/VISA/VISA Advanced/Bus-Interface Specific
Hope this helps!
-Ken
Oops. Just realized this was under the Labwindows CVI forum, not LabView. Anyway, it will help some.