Hi,
The limitation on the baud rate option is at the hardware level. Check this knowledge base article, which describes how the allowed baud rates for NI hardware is calculated, but it applies to any serial hardware: Which Baud Rates Do the National Instruments Serial Boards Support?
DiegoF
National Instruments
03-19-2007 01:26 PM
03-20-2007
08:01 AM
- last edited on
03-13-2025
02:15 PM
by
Content Cleaner
Message Edited by mvr on 03-20-2007 08:01 AM
03-20-2007 08:24 AM
03-20-2007 10:57 AM
The NI documentation is a little cryptic but it states:
"All NI serial hardware supports the standard baud rates. In addition, the PCI/PXI-843x family of hardware supports any baud rate between 2 bit/s and the maximum supported baud rate for that interface."
The maximum baud rate is 1000Kbits/s. They do not specify the resolution specifically but it was listed in the data sheet as at least as good as 1% which is well beyond the 3-5% required by a typical application. Unless you have a really extreme timing accuracy requirement (like testing a serial port at the timing boundry conditions) it should easily support the baud rate you specified. You may want to call or email your NI sales rep or app engineer to verify the details.