weird thermocouple behaviour with CVI and timer

I am using LW/CVI 4.0 and am performing DAQ with an SCXI 2000 chassis containing
SCXI 1200, 1102, 1161, 1303. We are taking Thermocouple readings each second
or so and send an output signal based upon these readings. The program runs
a timer loop while in operation. It works fine most of the time, but every
so often the temperature readings, which usually measure between 200 and
500 F, will start to drop, first to about 50, then they become negative and
march quickly off towards negative infinity, decreasing greatly each time
through the loop. Other times they will march upwards, becoming large positive
numbers quickly. However we know the actual temperature has not deviated
from between 200 and 500 degre
es. Does anyone know what on earth could be
causing this? Could it be electrical noise? I cannot figure this out for
the life of me.
Can one open thermocouple cause the rest of them to be misread? If so is
there a function I can use to get around this? Any help would be gratefully
appreciated.
Jeff Blackburn
0 Kudos
Message 1 of 3
(3,023 Views)

i am facing such problem! who can tell me what is the matter?

if the thermocouple is floating , all is ok;while i connect the thermocouple to the earth, it will swing greatly!

0 Kudos
Message 2 of 3
(2,903 Views)

The original problem experienced by buzzard sounds like it may have been an open connection.  Your issue sounds a lot like 60Hz noise caused by a grounding issue.  Take a look at this link  http://forums.ni.com/ni/board/message?board.id=300&message.id=934&requireLogin=False.  You might also want to search the Signal Conditioning forum.

 