PXI-2575 minimum load

I work on new project. I will use multimeter PXI-4070 with switch mux PXI-2575. The multimeter will measure voltage on differents tests points of a circuit in test. The input impedance of PXI-4070 are 10Mohms in measure voltage mode. So the current in relays of PXI-2575 will be very low <2.4uA. All voltages to measure are between 0.5VDC et 24VDC. There are also a measure of 8VaC and 16VAC.


The relay minimum load in datasheet of PXI2575 FEV2008 are "10uA at 10uV DC".
The relay minimum load in datasheet of PXI2575 APR2009 are "20mV/1mA".


My questions:

What the minimum load change in 2009 or it's a error in documentation?

What the real signification of "20mV/1mA"?

The current must be always greater than 1mA or periodicly greater than 1mA?

The voltage are voltage to init contact or the voltage must be always greater than 20mV?

What a current less than 1mA decrease life of relay?

If I measure AC voltage what the consequence on the measure when the current are between -1mA and 1mA?

 

Thank you!

0 Kudos
Message 1 of 2
(6,136 Views)

How is this related to the Excel add-on program called 'Measure'. I think you should post to the multi-function DAQ board.

0 Kudos
Message 2 of 2
(6,132 Views)