Triggerring PXI-4110 to measure 1 current value while HSDIO PXI-6552 generating waveform

Hi,

 

Some question about PXI-4110 to measure current while PXI-6552 is generating the waveform. 

 

1. Let say, I need to measure 3 points of current values, i.e. while PXI-6552 is generating sample-1000, 2000 and 3500. On the edge of sample 1000,2000 and 3500, the PXI-6552 will send a pulse via PFI line or via PXI backplane trigger line. My question is, is it possible to trigger PXI-4110 (hardware trigger or software trigger) to measure current values at these points ?

 

2. Let say I need to measure the current on 0ms (start of waveform generation by PXI-6552) , 1ms, 2ms, 3ms, 4ms... and so on for 1000 points of measurement, code diagram as shown at the figure below. It is possible for the VI "niDCPower Measure Multiple" to measure exactly at 1ms, 2ms, 3ms .. ? How much time will have to spend to complete acquire 1 point of measurement by "niDCPower Measure Multiple" ?

 

 

 

 

Thanks for viewing this post. Your advice on hardware used or software method is much appreciated. Thanks in advance.  

 

Message Edited by engwei on 02-02-2009 04:24 AM
0 Kudos
Message 1 of 3
(7,013 Views)

Hi engwei,

1. Unfortunately, the 4110 does not support hardware triggering. Therefore you cannot implement direct triggering through the backplane or anything like that. However, there are a couple of possible workarounds you can try:

 

a) Use software triggering: Say your 6552 is generating in one while loop, and your 4110 is to measure in another while loop. You can use a software syncrhonization method like notifiers to send a notification to your 4110 loop when your 6552 has generated the desired sample. This method, however, will not be very deterministic because the delay between the trigger and the response depends on your processor speed and load. Therefore, if you have other applications running in the background (like antivirus) it will increase the delay.

 

b) Use hardware triggering on another device: If you have another device that supports hardware triggering (like maybe an M-series multifunction DAQ module), you can configure this device to be triggered by a signal from the 6552, perform a very quick task (like a very short finite acquisition) then immediately execute the DCPower VI to perform the measurement. The trigger can be configured to be re-triggerable for multiple usage. This will most likely have a smaller time delay then the first option, but there will still be a delay (the time it takes to perform the short finite acquisiton on the M-series). Please refer to the attached screenshot for an idea of how to implement this.

 

2. To make your 4110 measure at specific time intervals, you can use one of the methods discussed above. As for how long it will take to acquire 1 measurement point, you may find this link helpful: https://www.ni.com/en/support/documentation/supplemental/08/source-and-measurement-timing-for-the-px...

 

This article is meant for the PXI-4130 but the 4110 has the same maximum sampling rate (3 kHz) and so the section discussing the speed should apply for both devices.

 

Under the Software Measurement Rate section, it is stated that the default behavior of the VI is to take an average of 10 samples. This corresponds to a maximum sampling rate of 300 samples/second. However, if you configure it to not do averaging (take only 1 sample) then the maximum rate of 3000 samples/second can be achieved.

 

It is also important to note that your program can only achieve this maximum sampling rate if your software loop takes less time to execute than the actual physical sampling. For example, if you want to sample at 3000 samples/second, that means that taking one sample takes 1/3000 seconds or 333 microseconds. If you software execution time is less than 333 microseconds, then you can achieve this maximum rate (because the speed is limited by the hardware, not the software). However, if your software takes more than 333 microseconds to execute, then the software loop time will define the maximum sampling rate you can get, which will be lower than 3000 samples/second.

 

I hope this answers your question.

 

Best regards,
Vern Yew
Applications Engineer, NI ASEAN

Best regards,
Vern Yew
Applications Engineer
0 Kudos
Message 2 of 3
(6,900 Views)

Hi Vern Yew,

 

Ya, this answer my question. Thanks for your very-detail reply.

 

BR,

engwei